Background information from one who  participated in the development of the equivalence factors for the  Navy. 

CLEANING SMT February 1998   



IONIC TESTING: A  PRIMER. 

By William G. Kenyon

Industry users continue to ask about the ionic test  method that was first published in MIL-P-28809 and now appears in the  J-STD-001B.  As one of the people  involved in the development of the equivalence factors, I decided to write this  column as a “public service” for both experienced and new members of our  industry. 

T.  F. Egan  of Bell Labs was looking for a way to determine if gold plated parts were clean  enough for spot welding.  He devised  a test method that immersed a given number of such parts in deionized water,  measuring the conductivity of the water before and after the parts were  introduced.  Because the parts were  identical, limiting controlling the number of parts controlled the surface area  exposed to the DI water.  Since this  test was only concerned with plating salt residues, the use of DI water,  pre-equilibrated with air to eliminate the effect of carbon dioxide, was a good  choice.  Egan looked at the delta C  in the conductivity before and after the test.  If there was less than 1.0 micromho/cm  increase for an area of 5 sq.  in.,  the parts were clean enough to go to the production line.  If not, they were re-rinsed until they passed, plus the rinsing steps in the plating line were checked to make sure  

they met process standards, with corrective action taken as needed.   

In the early 1970s, the military was encountering  numerous failures in radios and other conformally coated electronic hardware  deployed in Southeast Asia.  After investigation, the Navy lab in  Indianapolis  found the cause to be the ionic residues under the conformal coatings.  These residues would react with the  water vapor that permeated the coatings, which turned to liquid if  

conditions dropped below the dew point.   The ionics, dissolved in liquid  

water, could conduct current under the coating to cause the failures.  Since the organic conformal coatings are  all permeable to water vapor, the key was to eliminate, or significantly reduce,  the ionic residues.  The various  phases of this project were published as Material Research Reports (MRR).  First, the flux residue had to be  dissolved away from the assembly surface so the ionics 

could be measured.  The Navy   

group tested a number of 2-propanol/deionized water  mixtures for their ability to dissolve the solvent cleaned rosin flux  residues found on military hardware of that era.  The best solution was a 75/25  volume/volume 2-propanol/deionized water mixture.  Next, the team had to determine how to  wash the test boards and what pass/fail limits should be used.  So this phase was based on the Egan  test.  Thus a given volume of test  fluid was applied via a wash bottle per given area of PWA surface (10  milliliters per square inch).  The solution was purified by passage through a resin bed so the resistivity was  higher than 6 megohm-cm.  After the  washings had been collected in a clean vessel, the final resistivity was  measured.  It had to be above 2  megohm-cm.  

Titration of the test  solution with standardized NaCl solution showed that ionics equivalent to 1.26  micrograms of NaCl/sq.  cm.  of PWA surface would drop the resistivity from 6 to 2 megohm-cm.  The starting and final resistivity numbers were developed using the  levels on PWAs that performed satisfactorily in field service.  At this point the Navy had developed a  standardized test solution to remove a known flux residue, plus a test method to  measure the ionics and a pass/fail criteria based on field data (MRR 3-72).  The method issued in MIL-P-28809; Printed Wiring Assemblies, March, 1975.  (The spec first issued as MIL-P-XXXXX or  “MIL-P-Quint X”, to use the verbal shorthand of the

time.)

The flux makers developed commercial instruments for  automating the cleanliness test found in MIL-P-28809.  First the Ionograph™, providing data in  terms 

of conductivity, then the Omegameter™, using resistivity.   At that time, all 

military work  was done within the military specification system, so any change had to be  authorized by a waiver or deviation.  The military allowed a single waiver for use of the Ionograph™ (see MRR  5-75) instead of the wash bottle technique, then later a corresponding waiver  was developed and issued for the Omegameter™ (see MRR 10-76).  Other military contractors wanting to  use these instruments would then cite this documentation in applying for their waiver.  The paperwork for waivers  soon got so out of hand that the Navy lab in Indianapolis had to develop a method to  eliminate the waivers, thus allowing contractors to use any of several methods  to comply with the cleanliness test requirements in their  contracts. 

This was the beginning of the equivalence factor work,  which has caused a great deal on confusion.  At the time, the Navy, realizing that  component area was quite significant, developed component area measurements for  all the components used for military work.  The user would then just measure the length and width of the PWB,  multiply to get the area of one side, double it to get the area of both sides  and then add the sum of the component areas to get the full surface area of the  PWA to be tested.  The Navy lab  developed a series of test boards and two lab fluxes that would simulate RMA and  RA fluxes for use in a test plan to qualify the various commercial test  instruments for general use.  An  expert group, including the inventors of the commercial instruments, was invited  to Indianapolis  to help review and interpret the resulting data.  The activator level in the RMA flux was  really in the RA range, so we were faced with a single data set instead of the  planned two sets.  The instruments  included two resistivity/conductivity bridges (Beckman and Markson), the  Ionograph™, the Omegameter™ and the non-commercial Ion Chaser.  Since the military assemblies of the  time included examples of all the different components, we pooled the data for  each instrument to determine the average contamination level for all test boards  for a given instrument.  Next, the  averages for the two bridges were averaged, so we now had four averages.  The average for the bridges was chosen  as the standard, since the bridge was used as the original measurement  instrument.  The data sets were then  normalized by dividing the average for each instrument by the grand average for  the bridges.  

This gave the  “equivalence factors” for each instrument, with the bridges set 

equal to  1.0.   Next, each equivalence  factor was multiplied by 1.56 

micrograms NaCl/sq.  cm.  to give the pass/fail limit for that  instrument equal to the Mil-P-28809 limit of 2 megohm-cm.  Then, the resulting numbers were multiplied by 6.452 to convert them to the values needed to express the  pass/fail limit or instrument acceptance limit” when the area was measured in  square inches. (The higher 1.56 micrograms NaCl/sq.  cm.  number instead of the original 1.28 came  into use when I modified the electronics in an OmegaMeter 300 to give a 3X scale  expansion on the digital resistivity meter.  This showed that the clean solution was  really much higher than the 6 megohm-cm used to get the 1.28 micrograms  NaCl/sq.  cm.  value.) The details of the equivalence factor study can be found  in MRR 3-78.  The report concludes:  “All of the  instruments tested are suitable for determining ionic contaminants remaining  after solvent cleaning of post-soldering rosin-flux residues as described in  MIL-P-28809 providing the interim equivalence factors are employed.”  Meanwhile, W. B. Wargotz of Bell  Laboratories had been studying the deleterious effect of NaCl residues on  telecom boards and subjecting them to the 40 year life test.  Wargotz reported at the Fall,  78 IPC meeting that levels of less than  1 microgram NaCl/cm2 passed the 40 year life test, while levels of 7  micrograms NaCl/cm2 or higher would fail.  The 1-7 micrograms NaCl/cm2  range was a grey area.  Above 7 was a guaranteed failure area. However, the correlation of this  laboratory data with the empirical Navy data validated the pass/fail limit  established in MRR 3-72 and confirmed in 

MRR 3-78.   

At this point the industry had a valid test method for  measuring ionics from solvent  cleaned post-soldering rosin-flux residues.  The question of validity of the interim  equivalence factors was addressed by an experimental study conducted by the  makers of the Omegameter™.  The  study concluded the equivalence factors were applicable to newer models of this  instrument.  At this 

point the  equivalence factors became fixed.   

Northern Telecom investigated the validity of using  ionic testing for water-soluble fluxes.  Their findings, presented at the IPC by NT’s T. Duyck in a report on  their Deltameter, showed a dramatic difference in release rate of the residues  of rosin fluxes and water soluble fluxes from the surface of the 

test  boards.   In the typical 10-15  minute test period, rosin flux residues 

were about 90% percent extracted, while  it took about two hours to achieve the same level of extraction with  water-soluble fluxes.  Thus, using  the standard ionic test method with water-soluble flux residues gave numbers  that were very low, but not representative of the residues that were really  present.  

Synthetic Activated (SA)  fluxes were tested for release rate and found to be as fast or faster than rosin  fluxes.  The standard method was  shown to be applicable to SA and rosin flux residues, but not to water-soluble  flux residues. 

 Later, when the  ozone depleting solvents used in the electronics industry were phased out, the  IPC Cleaning & Coating Committee re-examined ionic testing to determine the  proper test solution and pass/fail limit appropriate for residues from  alternative soldering/cleaning processes, including no clean.  

The reports of this task group are  available from IPC.   

The ionic test is still valid today for rosin flux  residues, provided the equivalence factors are used (J-STD-001C).  Since many users employ the test to comply with telecom requirements or to monitor their processes, the applicability to modern water-soluble and no clean flux residues needs to be established.  Investigations carried  out by IPC members showed that high loadings of adipic acid, a major ingredient  in most no clean fluxes, had little deleterious effect on surface insulation  resistance.  The ionic test is  difficult to use in this case, since adipic acid is not very soluble in typical  solvents or in water.  Advances in  water-soluble flux formulations to eliminate the polyethylene glycols that  degrade SIR provide us with a new set of residues that need to be assessed.  The ionic test has been such a valued  process monitoring tool in addition to providing reliable electronics that the  industry needs to press for the research needed to keep it current with today’s  soldering and cleaning technologies. 





[Had many requests for the full text of Feb98 column vs  the shortened version actually published, here is what was cut from that Feb.  col.] 



IONIC TESTING: PART  2. 

By: William G.  Kenyon

Many readers have asked for more information on the  ionic testing primer provided in the February column.  I have expanded the text that was cut  from that column for space reasons, so between the two columns, a fairly  complete account is now available.  I will also try to answer some of the questions that have come up in the  intervening months. 

Q.  Why do we use 75/25 alcohol/water? 

A.  The Navy Lab that developed the test looked carefully at the test solution,  testing 25/75, 50/50 and 75/25 alcohol/water mixtures.  They found the 50/50 and 25/75 mixtures  did not dissolve all the rosin matrix, thus not all the ionics were released  into the solution for measurement (MRR 5-77).  This is the reason 75/25 was selected as  the official test solution. 

Q.  Why not 50/50 alcohol/water, since it is  alleged to be “more sensitive”

? 

A.  Some workers, ignoring Q&A #1, looked at the larger numbers obtained with  50/50 compared with 75/25, and concluded it was “more sensitive.” In reality,  they were seeing the greater dissociation of incompletely dissociated ionic  materials.  In other words,  compounds like table salt are completely dissociated into ions as long as there  is minimum water present, while incompletely dissociated flux ingredients  dissociate into measurable ions depending on the amount of water present.  The more water in the test solution, the  greater the dissociation, and the higher the number; thus leading to the claim  of “ greater sensitivity”.  

Q. How  do you comply with MIL-P-28809 with a 50/50 test  solution? 

A.  Electronics makers in Europe often used ionic  test equipment designed to run on 50/50 alcohol/water, which gave significant  problems when trying to comply with MIL-P-28809 limits designed around  75/25.  Since the values from identical test boards run in 50/50 could be as much as 10x higher than the values from 75/25, some electronics makers were desperate when achieving of  only ” 2 micrograms/sq.  cm., which in real numbers could be as  low as 0.2 micrograms/sq.  cm., when the pass/fail limit was  1.56  micrograms/sq.  cm.  Thus many electronics makers were either  over-cleaning or discarding valuable 

electronics hardware that was really quite  adequately cleaned.   

[In Europe, the  Contaminometer was used with 50/50 (ignoring the above) to compensate for the  rather insensitive Myron L conductivity probe used in it. 

The probe should be  0.01 but the Myron L had a cheap 0.1 probe which was fine for plating baths, but  not for this work.] Besides the promotion of the 50/50 as “more sensitive”,  there were also problems with some of the early versions of the test equipment,  especially the test cells.  Initially, many of these were made of acrylate plastic and the corresponding adhesives, which were not compatible with the more aggressive

75/25 v/v test solution.  Once this  compatibility problem was recognized, the instrument makers quickly converted to  other plastics and the problem disappeared. [Note: MIL-P-28809 has been  canceled.  The ionic cleanliness requirements can now be found in ANSI  J-STD-001.] Q. Some  instruments use conductivity and some use resistivity.  Why? 

A. Conductivity,  which is linear, is the reciprocal of resistivity, which is logarithmic.  The Ionograph™ used conductivity, which  was also used by T. 

Egan at Bell Labs.  The advantage of conductivity was that a recent reading that was twice  what an earlier reading had been meant that your PWAs had twice as much ionic  residue on them. Resistivity values were not as easy to interpret, since we  don’t tend to think in powers of ten.  The Omegameter™, which used resistivity, did so in order to automate the  original Navy manual ionic testing procedure. Initially this caused much  confusion in the industry, until enough papers were published that the industry  was comfortable with the two systems.  

Q.  I’ve got an older Omegameter™ that uses a little slide rule to generate the  actual ionic cleanliness readings.  The meter on the machine goes up to 20 megohm-cm, while the numbers on  the slide rule go up to 30 megohm-cm.  Can I just modify the circuitry of my instrument to take advantage of the  higher reading capability of the slide rule? 

A.  A careful check of the slide rule will show that you can paste in some log paper  (if you can find it) to continue the log scale to higher values.  

However, the electronics in the read-out  module of your instrument won’t track the log scale graph paper above 30  megohm-cm.  If you are planning on  doing a scale expansion from 20 megohm-cm to 60 megohm-cm, then it is best to  attach a precision resistance box to the terminals on the back of the instrument where the resistivity probe is usually attached.  Then track the actual resistance against  the readings to generate a graph of observed readings vs. actual 

values.   

[This  log paper patch worked quite well with my Omegameter™ 300.  We did a 3x scale expansion on it as  described, I still have a copy of the report with the circuit design, which was  about $200 to build and install. The precision resistor box is not included in  the price.  The electronics in the Omegameter weren’t linear above about 30 megohm-cm. ] Q.  In the February article, you explained “equivalence factors”. Is there a  tabulation of these factors available? 

A.  The equivalence factors are summarized in this table, taken from MRR 3-78, which  concludes: “All of the instruments tested are suitable for determining ionic  contaminants remaining after solvent cleaning of post-soldering rosin-flux  residues as described in MIL-P-28809 providing the interim equivalence factors  are employed.” 

Instrument

Average  NaCl/in2

Equivalence  Factor

Acceptance

Limit, micro-

grams

NaCl/cm2  

Acceptance Limit, micrograms   

NaCl/in2   

Beckman  bridge

7.47

1.56  

10.06   

Markson  bridge

7.62

1.56  

10.06   

Combined  bridges

7.545

1.0

1.56  

10.06   

Ionograph™

15.20

2.01

3.1  

20   

Omegameter™

10.51

1.39

2.2  

14   

Ion  Chaser

24.50

3.25

5.1

33

Loose Ends

The Department of Defense, realizing that this work  needed to be extended to cover saponifier cleaned rosin flux residues plus water  cleaned water soluble flux, fusing fluid and hot air leveling fluid fluxes,  proposed new studies to determine the proper test solutions and pass/fail limits  for the residues from these technologies.  This work was never funded and thus never done (see 

Appendix 2, DoD  Soldering Standardization Study).    

[As I recall, this was published in 1978-1980  timeframe.] W G. KENYON, an SMT Editorial Advisory Board member, is  president of Global Centre Consulting,  3336 Birmingham Drive, Fort Collins, CO 80526-2336. Tel: 

(970)  207-9586;  Cell: (970)  980‑6373;email [log in to unmask] 

(mailto:[log in to unmask])    

PULL QUOTES

The ionic  test is still valid for rosin and SA flux residues, provided the  

equivalence factors are used.”                            --End--



---------------------------------------------------

Technet Mail List provided as a service by IPC using LISTSERV 1.8e To unsubscribe, send a message to [log in to unmask] with following text in the BODY (NOT the subject field): SIGNOFF Technet To temporarily halt or (re-start) delivery of Technet send e-mail to [log in to unmask]: SET Technet NOMAIL or (MAIL) To receive ONE mailing per day of all the posts: send e-mail to [log in to unmask]: SET Technet Digest Search the archives of previous posts at: http://listserv.ipc.org/archives Please visit IPC web site http://www.ipc.org/contentpage.asp?Pageid=4.3.16 for additional information, or contact Keach Sasamori at [log in to unmask] or 847-615-7100 ext.2815

-----------------------------------------------------




ATOM RSS1 RSS2